Anaesthesia · Intravenous and inhalational agents: methylene blue and pulse oximetry
Sample itemAfter a methylene blue infusion for refractory vasoplegia, the pulse oximeter reading drops suddenly from 98 to 82 percent, yet the patient is pink and stable and an arterial blood gas shows a PaO2 of 280 mmHg. What explains the reading?
Original practice item, not a recalled exam question. Nothing here is recorded.
The answer and the reasoning
D. Methylene blue absorbs light at the oximeter wavelengths, giving a spuriously low SpO2
Intravenous dyes such as methylene blue, indocyanine green and indigo carmine absorb light at the oximeter wavelengths and cause a transient falsely low saturation. A normal PaO2 on blood gas analysis confirms the reading is artefactual.
Option by option
A. Interference by fetal haemoglobin
Fetal haemoglobin is irrelevant in an adult.
B. Severe intrapulmonary shunting not detected by the blood gas
A shunt large enough to produce this fall would be reflected in a low PaO2.
C. Carboxyhaemoglobin formation that reads as deoxygenated haemoglobin
Carboxyhaemoglobin is read as oxyhaemoglobin, and it was not formed here.
D. Methylene blue absorbs light at the oximeter wavelengths, giving a spuriously low SpO2 (correct)
Correct. The dye interferes with the optical measurement, and the blood gas excludes real hypoxaemia.
Source note: Original NextPrep sample item — pending expert review. This item is original practice material awaiting independent expert review, so confirm important facts in a standard textbook.
More Anaesthesia questions
- Intravenous anaestheticsWhich intravenous anaesthetic agent typically increases heart rate and blood pressure?
- Regional anaesthesiaWhat is the definitive treatment for persistent post-dural puncture headache that does n…
- Inhalational anaestheticsWhich of the following inhalational agents has the lowest MAC, and so the highest potency?
- Pre-operative assessmentA patient with well-controlled mild hypertension and no functional limitation is classif…