Radiology · Ultrasound: Doppler angle
Sample itemDuring a carotid Doppler examination, no flow signal is detected even though the vessel is patent. Which technical factor best explains this?
Original practice item, not a recalled exam question. Nothing here is recorded.
The answer and the reasoning
B. The ultrasound beam is at 90 degrees to the direction of flow
Doppler shift is proportional to the cosine of the angle between beam and flow, and cos 90 is zero, so no shift is recorded. An angle of 60 degrees or less is recommended for accurate velocity measurements. Angle correction is applied to the spectral trace.
Option by option
A. Gel has been applied
Gel is needed to remove air and does not stop signals.
B. The ultrasound beam is at 90 degrees to the direction of flow (correct)
Correct. A perpendicular beam gives no Doppler shift.
C. The beam is at 60 degrees to the flow
At 60 degrees a signal is recorded and velocity can be calculated.
D. The frequency is too high for a superficial vessel
High frequency does not abolish a signal in a superficial vessel.
Source note: Original NextPrep sample item — pending expert review. This item is original practice material awaiting independent expert review, so confirm important facts in a standard textbook.