Sample itemA sonographer finds no Doppler signal from a vessel lying exactly perpendicular to the ultrasound beam, even though blood is flowing. Why?
Original practice item, not a recalled exam question. Nothing here is recorded.
The answer and the reasoning
D. The Doppler shift depends on the cosine of the angle between beam and flow, which is zero at 90 degrees
The Doppler frequency shift is proportional to the cosine of the angle of insonation. At 90 degrees the cosine is zero, so no shift is detected. Angles below about 60 degrees are preferred for accurate velocity measurement.
Option by option
A. Aliasing occurs only when the beam is perpendicular to flow
Aliasing arises when the velocity exceeds what the pulse repetition frequency can measure, whatever the angle.
B. The beam is reflected completely and cannot enter the vessel
The beam penetrates the vessel normally, and the problem is the lack of a frequency shift.
C. Red cells move too slowly when flow is at right angles to the beam
Red cell speed does not depend on the angle of the probe.
D. The Doppler shift depends on the cosine of the angle between beam and flow, which is zero at 90 degrees (correct)
Correct. At a right angle, none of the flow velocity is directed along the beam.
Source note: Original NextPrep sample item — pending expert review. This item is original practice material awaiting independent expert review, so confirm important facts in a standard textbook.
More Radiology questions
- Ultrasound physicsWhat causes posterior acoustic shadowing?
- Ultrasound physicsWhat enhancement occurs behind a simple cyst?
- NeuroimagingWhat is the first investigation of choice in suspected acute subarachnoid haemorrhage?
- Cardiac imagingA 'boot-shaped' heart (coeur en sabot) on a chest radiograph is typical of: